  1. copy of letter received from BD:

    January 29, 2004

    Re: Important Product Recall

    Dear Valued Customer,

    This letter is to inform you that test strips from certain lots of BD Test Strips may produce an
    increased frequency of E-3 messages when used during blood glucose testing. The E-3 message is a
    protective feature of the meter that appears during a test to indicate issues with the application
    of a blood sample or with the test strip. It is important to note that this issue does not cause
    erroneous test results as no test result is shown on the meter's screen when the E-3 message is
    displayed. To maintain the highest product quality standards and to ensure your continued
    satisfaction, BD is performing this voluntary product recall and is replacing strips from all
    affected lots. This letter defines the affected lots and the process for product replacement.

    Replacement Process: BD requires that you follow these steps to determine if you need to replace any
    test strip product.

    1. Please locate the lot number on your BD Test Strip vial or box: just below the Code number on the
    vial and on the side of the box.

    2. Compare your test strip lot numbers to the following list: Affected BD Test Strip Lot Numbers
    2064322 3064016 3064090 3064156 3064279 2064333 3064018 3064104 3064167 3071009 2071311 3064030
    3064107 3064170 3071153 2078294 3064069 3064118 3064182 3078128 2120336 3064076 3064121 3064188
    3092028 3057009 3064079 3064125 3064237 3106007 3064013 3064083 3064132 3064145

    3. If you "do not" have any BD test strip vials or boxes with the above lot numbers there is no need
    to contact BD on the matter. You are welcome to call us at 1-866-556-8123 should you have any
    questions about your BD products.

    4. If you "do" have BD Test Strip vials or boxes with the above lot numbers we would like to replace
    them at no cost to you. Please call us toll free at 1-866-556-8123 (Monday through Friday, 8 am -
    midnight EST, and Saturday and Sunday between 8 am - 8 pm EST) so we can quickly send replacement
    test strips to you.

    While waiting to receive your replacement product you can continue to use your BD Test Strips.
    However, you may obtain a higher rate of E-3 messages instead of a test result. Your replacement
    product will correct this issue. If you continue to use test strips from the affected strip lots,
    ensure that you have an adequate supply with you.

    We Apologize for this inconvenience and thank you for your continued support of BD products.

    Sincerely,

    W. Marshall President BD Medical, Diabetes Care

    Mack Type 1 since 1975 alt-support-diabetes.orgalt-support-diabetes.org insulin-pumpers.orginsulin-pumpers.org

    In tribute to the United States of America and the State of Israel, two bastions of strength in a
    world filled with strife and terrorism.

    "To announce that there must be no criticism of the
    President, or that we are to stand by the President
    right or wrong, is not only unpatriotic and servile,
    but is morally treasonable to the American public."

    ...Theodore Roosevelt

  2. Thanks for posting the recall letter. 🙂 I'd like to reiterate my thoughts on the subject. The
    letter states that the lots mentioned are the ones causing the problems. In fact, none of the strips
    that BD has ever furnished (including those not referenced in the recall and those handpicked as
    suitable replacements) are any good.

    FWIW, I have returned to them a 3 month supply for a refund, which they have indicated is on its
    way to me.

    If anyone is comfortable with an improvement from 80% failure to 20% failure, so be it. 20% waste on
    something like this is unacceptable, in my book. I have also found that the strips that do actually
    give readings do not give what I consider to be accurate, or at least are not comparable to the way
    I feel or to my other meters.

    I would strongly advise anyone considering an upgrade to 512 or 712 pump to discuss your options
    before wasting a portion of your money on that part of the package.

    Randy

    In misc.health.diabetes Mack said:

    copy of letter received from BD:

    Quoted message said:

    January 29, 2004

    Quoted message said:

    Re: Important Product Recall

    Quoted message said:

    Dear Valued Customer,

    Quoted message said:

    This letter is to inform you that test strips from certain lots of BD Test Strips may produce an
    increased frequency of E-3 messages when used during blood glucose testing. The E-3 message is a
    protective feature of the meter that appears during a test to indicate issues with the application
    of a blood sample or with the test strip. It is important to note that this issue does not cause
    erroneous test results as no test result is shown on the meter's screen when the E-3 message is
    displayed. To maintain the highest product quality standards and to ensure your continued
    satisfaction, BD is performing this voluntary product recall and is replacing strips from all
    affected lots. This letter defines the affected lots and the process for product replacement.

    Quoted message said:

    Replacement Process: BD requires that you follow these steps to determine if you need to replace
    any test strip product.

    Quoted message said:

    1. Please locate the lot number on your BD Test Strip vial or box: just below the Code number on
    the vial and on the side of the box.

    Quoted message said:

    2. Compare your test strip lot numbers to the following list: Affected BD Test Strip Lot Numbers
    2064322 3064016 3064090 3064156 3064279 2064333 3064018 3064104 3064167 3071009 2071311 3064030
    3064107 3064170 3071153 2078294 3064069 3064118 3064182 3078128 2120336 3064076 3064121 3064188
    3092028 3057009 3064079 3064125 3064237 3106007 3064013 3064083 3064132 3064145

    Quoted message said:

    3. If you "do not" have any BD test strip vials or boxes with the above lot numbers there is no
    need to contact BD on the matter. You are welcome to call us at 1-866-556-8123 should you have
    any questions about your BD products.

    Quoted message said:

    4. If you "do" have BD Test Strip vials or boxes with the above lot numbers we would like to
    replace them at no cost to you. Please call us toll free at 1-866-556-8123 (Monday through
    Friday, 8 am - midnight EST, and Saturday and Sunday between 8 am - 8 pm EST) so we can quickly
    send replacement test strips to you.

    Quoted message said:

    While waiting to receive your replacement product you can continue to use your BD Test Strips.
    However, you may obtain a higher rate of E-3 messages instead of a test result. Your replacement
    product will correct this issue. If you continue to use test strips from the affected strip lots,
    ensure that you have an adequate supply with you.

    Quoted message said:

    We Apologize for this inconvenience and thank you for your continued support of BD products.

    Quoted message said:

    Sincerely,

    Quoted message said:

    W. Marshall President BD Medical, Diabetes Care

    Quoted message said:

    Mack Type 1 since 1975 alt-support-diabetes.orgalt-support-diabetes.org insulin-pumpers.orginsulin-pumpers.org

    Quoted message said:

    In tribute to the United States of America and the State of Israel, two bastions of strength in a
    world filled with strife and terrorism.

    Quoted message said:

    "To announce that there must be no criticism of the
    President, or that we are to stand by the President
    right or wrong, is not only unpatriotic and servile,
    but is morally treasonable to the American public."

    Quoted message said:

    ...Theodore Roosevelt

  3. in misc.health.diabetes Randy Sigman said:
    Quoted message said:


    I would strongly advise anyone considering an upgrade to
    512 or 712 pump to discuss your options before wasting a
    portion=20 of your money on that part of the package.=20

    Having just bought a 712, and having discovered (big
    surprise to me, but apparently, reading the groups,
    everyone's having that surprise) that the= bd meter sucks, I
    called minimed to ask about returing the meter for credit.=
    Was told that first, the bd meter is bd's responsibility,
    and more = importantly, that minimed provides the meter at
    no additional charge as part of the = package. So much for
    refunds. And it occured to me that it's still needed to =
    download the pump to the computer, at least with the data
    cable minimed sent me that = connects to the test strip port
    (with great difficulty).

    Peter

  4. Peter W. Rowe said:


    On Wed, 18 Feb 2004 18:00:50 GMT, in misc.health.diabetes

    Randy Sigman said:
    Quoted message said:


    I would strongly advise anyone considering an upgrade to
    512 or 712 pump to discuss your options before wasting a
    portion of your money on that part of the package.

    Having just bought a 712, and having discovered (big
    surprise to me, but apparently, reading the groups,
    everyone's having that surprise) that the bd meter sucks,
    I called minimed to ask about returing the meter for
    credit. Was told that first, the bd meter is bd's
    responsibility, and more importantly, that minimed
    provides the meter at no additional charge as part of the
    package. So much for refunds. And it occured to me that
    it's still needed to download the pump to the computer, at
    least with the data cable minimed sent me that connects to
    the test strip port (with great difficulty).

    Peter

    Hi Peter, and welcome to misc.health.diabetes (that's where
    I'm reading you from).

    Minimed's gotten such bad reviews here in recent months I
    finally settled on the Animas. My IR1000 is sitting in its
    box right beside me, I have to wait 2 weeks for my
    training :-(

    Crappy customer service, billing years after a purchase,
    false waterproof claims and the problems others have had
    with the meter made my mind up while I was battling the
    insurance company.

    Animas (distributed by AutoControl Medical in Canada) threw
    in an upgrade to an IR1200 to seal the deal.

    Now if I could just get *on* the freakin' thing! I've been
    waiting over 10 months. I'm getting itchy-[censored]!

    Vicki

  5. In misc.health.diabetes Peter W. Rowe said:

    On Wed, 18 Feb 2004 18:00:50 GMT, in misc.health.diabetes
    Randy Sigman <[email hidden]> wrote:

    Quoted message said:
    Quoted message said:
    Quoted message said:


    I would strongly advise anyone considering an upgrade to
    512 or 712 pump to discuss your options before wasting a
    portion of your money on that part of the package.

    Quoted message said:

    Having just bought a 712, and having discovered (big
    surprise to me, but apparently, reading the groups,
    everyone's having that surprise) that the bd meter sucks,
    I called minimed to ask about returing the meter for
    credit. Was told that first, the bd meter is bd's
    responsibility, and more importantly, that minimed
    provides the meter at no additional charge as part of the
    package. So much for refunds. And it occured to me that
    it's still needed to download the pump to the computer, at
    least with the data cable minimed sent me that connects to
    the test strip port (with great difficulty).

    Hi Peter,

    That the meter is BD's responsibility is not something that
    we can really argue. That it came as part of a package,
    however, is. In fact, I was told quite the opposite of what
    you were told. I was told that the reason for the cost of
    the upgrade was largely because of the purchase that Minimed
    must make to the outside 3rd party (BD) for the meter. When
    I told them I wanted to return the entire upgrade, they said
    NO, you bought it, you own it. This, I thought, was reason
    for me to never consider purchasing product from them again.
    I took my concerns up to a higher level, and was eventually
    given a refund, and allowed to keep the goods.

    I never got any data cable from Minimed. I use my Glucometer
    Elite, for which I purchased a data cable. If I want to use
    the bolus wizard, I can simply enter the BG value obtained.
    It saves on the pump battery too. 🙂

    If you're interested in the email address of the party that
    ultimately helped me out, let me know, and I'll dig it up
    and post it.

    Best,

    Randy

  6. Hi Everyone:

    I've written letters to BD, Medtronic, and the Food and Drug
    Administration complaining about the lousy experience that I
    have had with BD's meter. BD has sent me replacement strips
    on two occassions and this last occassion, they sent me a
    new meter. I set up the meter with the time and date, stuck
    a strip in, and using control solution, obtained 5 E-3 error
    messages in a row. The 6th strip gave me a reading of 92. So
    I opened a brand new control solution bottle and proceeded
    to obtain 7 error messages in a row before I obtained a 128
    reading. I thought, give it the full treatment, stuck a new
    strip in, and got a blood sample. First strip was an E-3,
    stuck a new strip in and it read low. Took the new strips,
    new meter, and put it in the box along with the previous
    meter and strips and wrote BD a letter detailing what I
    personally thought of their crappy meter, and the fact that
    I had gone back to my old meter. I don't need Medtronics
    little wizard. I don't need BD's crappy little meter. And am
    quite satisfied using my old meter. Having been involved in
    Diabetes Camping for many years, I have had oportunities to
    try many different meters. Almost all of them do a pretty
    credible job of measuring blood glucose. Medtronic reallly
    screwed up badly picking BD to build a meter to go with the
    Paradym meter. They can deny all responsibility for this
    crummy glucose meter, but in the end they will be painted
    with the same brush. Becton Dickinson responded to my email
    complaint and explanation that I was returning to them both
    crummy meters. Their rep called me up on the phone. They
    didn't make excuses. Appologized for my troubles, and
    assured me that when they fixed the problems that they were
    having, would be happy to send me a new meter and strips.
    Too bad Medtronic wasn't as upfront as BD. BTW, I'm also
    using a Paradym 712 pump and a Lifescan Ultra Smart meter.
    The combination is working well for me. You have to do what
    you have to do to take care of yourself.

    Take care all, Bob

    "Randy Sigman" <[email hidden]> wrote in message
    "]news:[email hidden]...

    Quoted message said:

    In misc.health.diabetes Peter W. Rowe

    Quoted message said:

    On Wed, 18 Feb 2004 18:00:50 GMT, in
    misc.health.diabetes Randy Sigman
    <[email hidden]> wrote:

    Quoted message said:
    Quoted message said:

    >
    >I would strongly advise anyone considering an upgrade
    >to 512 or 712 pump to discuss your options before
    >wasting a portion of your money on that part of the
    >package.

    Quoted message said:

    Having just bought a 712, and having discovered (big
    surprise to me, but apparently, reading the groups,
    everyone's having that surprise) that


    the bd

    Quoted message said:
    Quoted message said:

    meter sucks, I called minimed to ask about returing the
    meter for


    credit. Was

    Quoted message said:
    Quoted message said:

    told that first, the bd meter is bd's responsibility,
    and more


    importantly, that

    Quoted message said:
    Quoted message said:

    minimed provides the meter at no additional charge as
    part of the


    package. So

    Quoted message said:
    Quoted message said:

    much for refunds. And it occured to me that it's still
    needed to


    download the

    Quoted message said:
    Quoted message said:

    pump to the computer, at least with the data cable
    minimed sent me that


    connects

    Quoted message said:
    Quoted message said:

    to the test strip port (with great difficulty).

    Hi Peter,

    That the meter is BD's responsibility is not something
    that we can really


    argue.

    Quoted message said:

    That it came as part of a package, however, is. In fact, I
    was told quite


    the

    Quoted message said:

    opposite of what you were told. I was told that the reason
    for the cost of


    the

    Quoted message said:

    upgrade was largely because of the purchase that Minimed
    must make to the


    outside

    Quoted message said:

    3rd party (BD) for the meter. When I told them I wanted to
    return the


    entire

    Quoted message said:

    upgrade, they said NO, you bought it, you own it. This, I
    thought, was


    reason

    Quoted message said:

    for me to never consider purchasing product from them
    again. I took my


    concerns

    Quoted message said:

    up to a higher level, and was eventually given a refund,
    and allowed to


    keep the

    Quoted message said:

    goods.

    I never got any data cable from Minimed. I use my
    Glucometer Elite, for


    which

    Quoted message said:

    I purchased a data cable. If I want to use the bolus
    wizard, I can simply


    enter

    Quoted message said:

    the BG value obtained. It saves on the pump battery too.
    🙂

    If you're interested in the email address of the party
    that ultimately


    helped

    Quoted message said:

    me out, let me know, and I'll dig it up and post it.

    Best,

    Randy

  7. abc said:

    Having been involved in Diabetes Camping for many years, I
    have had oportunities to try many different meters.

    I use the Ultra Smart BUT I find the Precision Xtra's
    lighting feature EXTREMELY valuable when I'm in a tent in
    the pitch dark.

  8. how about complaining to the higher ups at
    Medtronic/MiniMed? I could care less about BD, but I've got
    a vested interest in MiniMed doing the right thing. ('Course
    all this will be moot soon, as they will be coming out with
    their 515 in less than a year if all goes well and that will
    use the Guardian technology). So don't sweat the BD BS.

    dave

    abc said:

    Hi Everyone:

    I've written letters to BD, Medtronic, and the Food and
    Drug Administration complaining about the lousy experience
    that I have had with BD's meter. BD has sent me
    replacement strips on two occassions and this last
    occassion, they sent me a new meter. I set up the meter
    with the time and date, stuck a strip in, and using
    control solution, obtained 5 E-3 error messages in a row.
    The 6th strip gave me a reading of 92. So I opened a brand
    new control solution bottle and proceeded to obtain 7
    error messages in a row before I obtained a 128 reading. I
    thought, give it the full treatment, stuck a new strip in,
    and got a blood sample. First strip was an E-3, stuck a
    new strip in and it read low. Took the new strips, new
    meter, and put it in the box along with the previous meter
    and strips and wrote BD a letter detailing what I
    personally thought of their crappy meter, and the fact
    that I had gone back to my old meter. I don't need
    Medtronics little wizard. I don't need BD's crappy little
    meter. And am quite satisfied using my old meter. Having
    been involved in Diabetes Camping for many years, I have
    had oportunities to try many different meters. Almost all
    of them do a pretty credible job of measuring blood
    glucose. Medtronic reallly screwed up badly picking BD to
    build a meter to go with the Paradym meter. They can deny
    all responsibility for this crummy glucose meter, but in
    the end they will be painted with the same brush. Becton
    Dickinson responded to my email complaint and explanation
    that I was returning to them both crummy meters. Their rep
    called me up on the phone. They didn't make excuses.
    Appologized for my troubles, and assured me that when they
    fixed the problems that they were having, would be happy
    to send me a new meter and strips. Too bad Medtronic
    wasn't as upfront as BD. BTW, I'm also using a Paradym 712
    pump and a Lifescan Ultra Smart meter. The combination is
    working well for me. You have to do what you have to do to
    take care of yourself.

    Take care all, Bob

    "Randy Sigman" <[email hidden]> wrote in message
    "]news:[email hidden]...

    Quoted message said:

    In misc.health.diabetes Peter W. Rowe

    Quoted message said:

    On Wed, 18 Feb 2004 18:00:50 GMT, in misc.health.diabetes
    Randy Sigman <[email hidden]> wrote:

    Quoted message said:

    >>I would strongly advise anyone considering an upgrade
    >>to 512 or 712 pump to discuss your options before
    >>wasting a portion of your money on that part of the
    >>package.

    Quoted message said:

    Having just bought a 712, and having discovered (big
    surprise to me, but apparently, reading the groups,
    everyone's having that surprise) that

    the bd

    Quoted message said:
    Quoted message said:

    meter sucks, I called minimed to ask about returing the
    meter for

    credit. Was

    Quoted message said:
    Quoted message said:

    told that first, the bd meter is bd's responsibility,
    and more

    importantly, that

    Quoted message said:
    Quoted message said:

    minimed provides the meter at no additional charge as
    part of the

    package. So

    Quoted message said:
    Quoted message said:

    much for refunds. And it occured to me that it's still
    needed to

    download the

    Quoted message said:
    Quoted message said:

    pump to the computer, at least with the data cable
    minimed sent me that

    connects

    Quoted message said:
    Quoted message said:

    to the test strip port (with great difficulty).

    Hi Peter,

    That the meter is BD's responsibility is not something
    that we can really

    argue.

    Quoted message said:

    That it came as part of a package, however, is. In fact, I
    was told quite

    the

    Quoted message said:

    opposite of what you were told. I was told that the reason
    for the cost of

    the

    Quoted message said:

    upgrade was largely because of the purchase that Minimed
    must make to the

    outside

    Quoted message said:

    3rd party (BD) for the meter. When I told them I wanted to
    return the

    entire

    Quoted message said:

    upgrade, they said NO, you bought it, you own it. This, I
    thought, was

    reason

    Quoted message said:

    for me to never consider purchasing product from them
    again. I took my

    concerns

    Quoted message said:

    up to a higher level, and was eventually given a refund,
    and allowed to

    keep the

    Quoted message said:

    goods.

    I never got any data cable from Minimed. I use my
    Glucometer Elite, for

    which

    Quoted message said:

    I purchased a data cable. If I want to use the bolus
    wizard, I can simply

    enter

    Quoted message said:

    the BG value obtained. It saves on the pump battery too.
    🙂

    If you're interested in the email address of the party
    that ultimately

    helped

    Quoted message said:

    me out, let me know, and I'll dig it up and post it.

    Best,

    Randy


  9. aren't you trading accuracy for a light?? Earlier Precision
    meters were the absolute WORST meters on the market at one
    time. around '93-'95, IIRC. Did they play catch up with the
    market leader?

    dave

    Marshall said:
    abc said:

    Having been involved in Diabetes Camping for many years,
    I have had oportunities to try many different meters.

    I use the Ultra Smart BUT I find the Precision Xtra's
    lighting feature EXTREMELY valuable when I'm in a tent in
    the pitch dark.

  10. In misc.health.diabetes abc said:

    Hi Everyone:

    Quoted message said:

    I've written letters to BD, Medtronic, and the Food and
    Drug Administration complaining about the lousy experience
    that I have had with BD's meter. BD has sent me
    replacement strips on two occassions and this last
    occassion, they sent me a new meter. I set up the meter
    with the time and date, stuck a strip in, and using
    control solution, obtained 5 E-3 error messages in a row.
    The 6th strip gave me a reading of 92. So I opened a brand
    new control solution bottle and proceeded to obtain 7
    error messages in a row before I obtained a 128 reading. I
    thought, give it the full treatment, stuck a new strip in,
    and got a blood sample. First strip was an E-3, stuck a
    new strip in and it read low. Took the new strips, new
    meter, and put it in the box along with the previous meter
    and strips and wrote BD a letter detailing what I
    personally thought of their crappy meter, and the fact
    that I had gone back to my old meter. I don't need
    Medtronics little wizard. I don't need BD's crappy little
    meter. And am quite satisfied using my old meter. Having
    been involved in Diabetes Camping for many years, I have
    had oportunities to try many different meters. Almost all
    of them do a pretty credible job of measuring blood
    glucose. Medtronic reallly screwed up badly picking BD to
    build a meter to go with the Paradym meter. They can deny
    all responsibility for this crummy glucose meter, but in
    the end they will be painted with the same brush.

    I've told them that, but they aren't interested in hearing
    it. When they're existing customer base starts migrating to
    the competition, they might begin to wonder what is wrong,
    but I doubt even that will raise a flag.

    Quoted message said:

    Becton Dickinson responded to my email complaint and
    explanation that I was returning to them both crummy
    meters. Their rep called me up on the phone. They didn't
    make excuses. Appologized for my troubles, and assured me
    that when they fixed the problems that they were having,
    would be happy to send me a new meter and strips. Too bad
    Medtronic wasn't as upfront as BD.

    My experience with BD wasn't as pleasant. They insisted that
    their product was fixed, and that there was nothing left for
    them to do.

    BWahahahaha.

    Randy

  11. In misc.health.diabetes Bay Area Dave said:

    how about complaining to the higher ups at
    Medtronic/MiniMed? I could care less about BD, but I've
    got a vested interest in MiniMed doing the right thing.
    ('Course all this will be moot soon, as they will be
    coming out with their 515 in less than a year if all goes
    well and that will use the Guardian technology). So don't
    sweat the BD BS.

    You might want to reconsider your "vested interest"
    position. You and your wife may both currently have their
    product, but nothing is forcing you to "renew" your
    commitment to them. Additionally, whatever the 515 might
    have to offer, the service level, bad attitudes, and
    snottiness of their staff detracts from my future
    considerations of their offerings.

    Randy

    Quoted message said:

    abc wrote:

    Quoted message said:
    Quoted message said:

    Hi Everyone:

    I've written letters to BD, Medtronic, and the Food and
    Drug Administration complaining about the lousy
    experience that I have had with BD's meter. BD has sent
    me replacement strips on two occassions and this last
    occassion, they sent me a new meter. I set up the meter
    with the time and date, stuck a strip in, and using
    control solution, obtained 5 E-3 error messages in a row.
    The 6th strip gave me a reading of 92. So I opened a
    brand new control solution bottle and proceeded to obtain
    7 error messages in a row before I obtained a 128
    reading. I thought, give it the full treatment, stuck a
    new strip in, and got a blood sample. First strip was an
    E-3, stuck a new strip in and it read low. Took the new
    strips, new meter, and put it in the box along with the
    previous meter and strips and wrote BD a letter detailing
    what I personally thought of their crappy meter, and the
    fact that I had gone back to my old meter. I don't need
    Medtronics little wizard. I don't need BD's crappy little
    meter. And am quite satisfied using my old meter. Having
    been involved in Diabetes Camping for many years, I have
    had oportunities to try many different meters. Almost all
    of them do a pretty credible job of measuring blood
    glucose. Medtronic reallly screwed up badly picking BD to
    build a meter to go with the Paradym meter. They can deny
    all responsibility for this crummy glucose meter, but in
    the end they will be painted with the same brush. Becton
    Dickinson responded to my email complaint and explanation
    that I was returning to them both crummy meters. Their
    rep called me up on the phone. They didn't make excuses.
    Appologized for my troubles, and assured me that when
    they fixed the problems that they were having, would be
    happy to send me a new meter and strips. Too bad
    Medtronic wasn't as upfront as BD. BTW, I'm also using a
    Paradym 712 pump and a Lifescan Ultra Smart meter. The
    combination is working well for me. You have to do what
    you have to do to take care of yourself.

    Take care all, Bob

    "Randy Sigman" <[email hidden]> wrote in
    message "]news:[email hidden]...

    Quoted message said:

    In misc.health.diabetes Peter W. Rowe
    <[email hidden]> wrote:

    >On Wed, 18 Feb 2004 18:00:50 GMT, in
    >misc.health.diabetes Randy Sigman
    ><[email hidden]> wrote:

    >>>I would strongly advise anyone considering an upgrade
    >>>to 512 or 712 pump to discuss your options before
    >>>wasting a portion of your money on that part of the
    >>>package.

    >Having just bought a 712, and having discovered (big
    >surprise to me, but apparently, reading the groups,
    >everyone's having that surprise) that

    the bd

    Quoted message said:

    >meter sucks, I called minimed to ask about returing the
    >meter for

    credit. Was

    Quoted message said:

    >told that first, the bd meter is bd's responsibility,
    >and more

    importantly, that

    Quoted message said:

    >minimed provides the meter at no additional charge as
    >part of the

    package. So

    Quoted message said:

    >much for refunds. And it occured to me that it's still
    >needed to

    download the

    Quoted message said:

    >pump to the computer, at least with the data cable
    >minimed sent me that

    connects

    Quoted message said:

    >to the test strip port (with great difficulty).

    Hi Peter,

    That the meter is BD's responsibility is not something
    that we can really

    argue.

    Quoted message said:

    That it came as part of a package, however, is. In fact,
    I was told quite

    the

    Quoted message said:

    opposite of what you were told. I was told that the
    reason for the cost of

    the

    Quoted message said:

    upgrade was largely because of the purchase that Minimed
    must make to the

    outside

    Quoted message said:

    3rd party (BD) for the meter. When I told them I wanted
    to return the

    entire

    Quoted message said:

    upgrade, they said NO, you bought it, you own it. This, I
    thought, was

    reason

    Quoted message said:

    for me to never consider purchasing product from them
    again. I took my

    concerns

    Quoted message said:

    up to a higher level, and was eventually given a refund,
    and allowed to

    keep the

    Quoted message said:

    goods.

    I never got any data cable from Minimed. I use my
    Glucometer Elite, for

    which

    Quoted message said:

    I purchased a data cable. If I want to use the bolus
    wizard, I can simply

    enter

    Quoted message said:

    the BG value obtained. It saves on the pump battery
    too. 🙂

    If you're interested in the email address of the party
    that ultimately

    helped

    Quoted message said:

    me out, let me know, and I'll dig it up and post it.

    Best,

    Randy



  12. When my pharmacist sold me Ultra Strips for my Ultra Smart
    meter I did a test. I used Ultra Smart strips, a Precision
    Xtra machine and the Ultra strips. All results were pretty
    much within the same range, though I still do not understand
    why he keeps selling me the Ultra Strips, instead of the
    Ultra Smart Strips.

    Is there a difference?

    Bay Area Dave said:

    aren't you trading accuracy for a light?? Earlier
    Precision meters were the absolute WORST meters on the
    market at one time. around '93-'95, IIRC. Did they play
    catch up with the market leader?

    dave

    Marshall said:
    abc said:

    Having been involved in Diabetes Camping for many years,
    I have had oportunities to try many different meters.

    I use the Ultra Smart BUT I find the Precision Xtra's
    lighting feature EXTREMELY valuable when I'm in a tent in
    the pitch dark.


    --
    homepage.mac.commkatzman

  13. Randy Sigman said:

    In misc.health.diabetes Bay Area Dave <[email hidden]>

    Quoted message said:

    how about complaining to the higher ups at
    Medtronic/MiniMed? I could care less about BD, but I've
    got a vested interest in MiniMed doing the right thing.
    ('Course all this will be moot soon, as they will be
    coming out with their 515 in less than a year if all goes
    well and that will use the Guardian technology). So don't
    sweat the BD BS.

    You might want to reconsider your "vested interest"
    position. You and your wife may both currently have their
    product, but nothing is forcing you to "renew" your
    commitment to them. Additionally, whatever the 515 might
    have to offer, the service level, bad attitudes, and
    snottiness of their staff detracts from my future
    considerations of their offerings.

    Randy

    That depend on insurance. If their pumps are not
    malfunctioning and the warranties have run out, the
    insurance may choose not to cover the cost of switching to
    another pump. You'd have to read the policy.

    Mack©«®» Type 1 since 1975 alt-support-alt-support-
    diabetes.org insulin-pumpers.orginsulin-pumpers.org

    In tribute to the United States of America and the State of
    Israel, two bastions of strength in a world filled with
    strife and terrorism.

    "To announce that there must be no criticism of the
    President, or that we are to stand by the President
    right or wrong, is not only unpatriotic and servile,
    but is morally treasonable to the American public."

    ...Theodore Roosevelt

    Quoted message said:
    Quoted message said:

    abc wrote:

    Quoted message said:
    Quoted message said:

    Hi Everyone:

    I've written letters to BD, Medtronic, and the Food and
    Drug Administration complaining about the lousy
    experience that I have had with BD's meter. BD has sent
    me replacement strips on two occassions and this last
    occassion, they sent me a new meter. I set up the meter
    with the time and date, stuck a strip in, and using
    control solution, obtained 5 E-3 error messages in a
    row. The 6th strip gave me a reading of 92. So I opened
    a brand new control solution bottle and proceeded to
    obtain 7 error messages in a row before I obtained a 128
    reading. I thought, give it the full treatment, stuck a
    new strip in, and got a blood sample. First strip was an
    E-3, stuck a new strip in and it read low. Took the new
    strips, new meter, and put it in the box along with the
    previous meter and strips and wrote BD a letter
    detailing what I personally thought of their crappy
    meter, and the fact that I had gone back to my old
    meter. I don't need Medtronics little wizard. I don't
    need BD's crappy little meter. And am quite satisfied
    using my old meter. Having been involved in Diabetes
    Camping for many years, I have had oportunities to try
    many different meters. Almost all of them do a pretty
    credible job of measuring blood glucose. Medtronic
    reallly screwed up badly picking BD to build a meter to
    go with the Paradym meter. They can deny all
    responsibility for this crummy glucose meter, but in the
    end they will be painted with the same brush. Becton
    Dickinson responded to my email complaint and
    explanation that I was returning to them both crummy
    meters. Their rep called me up on the phone. They didn't
    make excuses. Appologized for my troubles, and assured
    me that when they fixed the problems that they were
    having, would be happy to send me a new meter and
    strips. Too bad Medtronic wasn't as upfront as BD. BTW,
    I'm also using a Paradym 712 pump and a Lifescan Ultra
    Smart meter. The combination is working well for me. You
    have to do what you have to do to take care of yourself.

    Take care all, Bob

    "Randy Sigman" <[email hidden]> wrote in
    message "]news:[email hidden]...

    >In misc.health.diabetes Peter W. Rowe
    ><[email hidden]> wrote:
    >
    >>On Wed, 18 Feb 2004 18:00:50 GMT, in
    >>misc.health.diabetes Randy Sigman
    >><[email hidden]> wrote:
    >
    >>>>I would strongly advise anyone considering an upgrade
    >>>>to 512 or 712 pump to discuss your options before
    >>>>wasting a portion of your money on that part of the
    >>>>package.
    >
    >>Having just bought a 712, and having discovered (big
    >>surprise to me, but apparently, reading the groups,
    >>everyone's having that surprise) that

    the bd

    >>meter sucks, I called minimed to ask about returing the
    >>meter for

    credit. Was

    >>told that first, the bd meter is bd's responsibility,
    >>and more

    importantly, that

    >>minimed provides the meter at no additional charge as
    >>part of the

    package. So

    >>much for refunds. And it occured to me that it's still
    >>needed to

    download the

    >>pump to the computer, at least with the data cable
    >>minimed sent me that

    connects

    >>to the test strip port (with great difficulty).
    >
    >Hi Peter,
    >
    >That the meter is BD's responsibility is not something
    >that we can really

    argue.

    >That it came as part of a package, however, is. In fact,
    >I was told quite

    the

    >opposite of what you were told. I was told that the
    >reason for the cost of

    the

    >upgrade was largely because of the purchase that Minimed
    >must make to the

    outside

    >3rd party (BD) for the meter. When I told them I wanted
    >to return the

    entire

    >upgrade, they said NO, you bought it, you own it. This,
    >I thought, was

    reason

    >for me to never consider purchasing product from them
    >again. I took my

    concerns

    >up to a higher level, and was eventually given a refund,
    >and allowed to

    keep the

    >goods.
    >
    >I never got any data cable from Minimed. I use my
    >Glucometer Elite, for

    which

    >I purchased a data cable. If I want to use the bolus
    >wizard, I can simply

    enter

    >the BG value obtained. It saves on the pump battery
    >too. 🙂
    >
    >If you're interested in the email address of the party
    >that ultimately

    helped

    >me out, let me know, and I'll dig it up and post it.
    >
    >Best,
    >
    >Randy
    >


  14. We will evaluate the other meters on the market, that's for
    sure. I was looking over the Cozmo the other day. My doctor
    has NEVER prescribed it and it looks pretty weird. Also, the
    boluses are given to fast for my taste. What else is
    available in the US that infuses boluses at no more than 2
    units per minute? Anything??

    dave

    Randy Sigman said:

    In misc.health.diabetes Bay Area Dave

    Quoted message said:

    how about complaining to the higher ups at
    Medtronic/MiniMed? I could care less about BD, but I've
    got a vested interest in MiniMed doing the right thing.
    ('Course all this will be moot soon, as they will be
    coming out with their 515 in less than a year if all goes
    well and that will use the Guardian technology). So don't
    sweat the BD BS.

    You might want to reconsider your "vested interest"
    position. You and your wife may both currently have their
    product, but nothing is forcing you to "renew" your
    commitment to them. Additionally, whatever the 515 might
    have to offer, the service level, bad attitudes, and
    snottiness of their staff detracts from my future
    considerations of their offerings.

    Randy

    Quoted message said:

    abc wrote:

    Quoted message said:
    Quoted message said:

    Hi Everyone:

    I've written letters to BD, Medtronic, and the Food and
    Drug Administration complaining about the lousy
    experience that I have had with BD's meter. BD has sent
    me replacement strips on two occassions and this last
    occassion, they sent me a new meter. I set up the meter
    with the time and date, stuck a strip in, and using
    control solution, obtained 5 E-3 error messages in a row.
    The 6th strip gave me a reading of 92. So I opened a
    brand new control solution bottle and proceeded to obtain
    7 error messages in a row before I obtained a 128
    reading. I thought, give it the full treatment, stuck a
    new strip in, and got a blood sample. First strip was an
    E-3, stuck a new strip in and it read low. Took the new
    strips, new meter, and put it in the box along with the
    previous meter and strips and wrote BD a letter detailing
    what I personally thought of their crappy meter, and the
    fact that I had gone back to my old meter. I don't need
    Medtronics little wizard. I don't need BD's crappy little
    meter. And am quite satisfied using my old meter. Having
    been involved in Diabetes Camping for many years, I have
    had oportunities to try many different meters. Almost all
    of them do a pretty credible job of measuring blood
    glucose. Medtronic reallly screwed up badly picking BD to
    build a meter to go with the Paradym meter. They can deny
    all responsibility for this crummy glucose meter, but in
    the end they will be painted with the same brush. Becton
    Dickinson responded to my email complaint and explanation
    that I was returning to them both crummy meters. Their
    rep called me up on the phone. They didn't make excuses.
    Appologized for my troubles, and assured me that when
    they fixed the problems that they were having, would be
    happy to send me a new meter and strips. Too bad
    Medtronic wasn't as upfront as BD. BTW, I'm also using a
    Paradym 712 pump and a Lifescan Ultra Smart meter. The
    combination is working well for me. You have to do what
    you have to do to take care of yourself.

    Take care all, Bob

    "Randy Sigman" <[email hidden]> wrote in
    message "]news:[email hidden]...

    >In misc.health.diabetes Peter W. Rowe
    ><[email hidden]> wrote:
    >
    >
    >>On Wed, 18 Feb 2004 18:00:50 GMT, in
    >>misc.health.diabetes Randy Sigman
    >><[email hidden]> wrote:
    >
    >>>>I would strongly advise anyone considering an upgrade
    >>>>to 512 or 712 pump to discuss your options before
    >>>>wasting a portion of your money on that part of the
    >>>>package.
    >
    >>Having just bought a 712, and having discovered (big
    >>surprise to me, but apparently, reading the groups,
    >>everyone's having that surprise) that

    the bd

    >>meter sucks, I called minimed to ask about returing the
    >>meter for

    credit. Was

    >>told that first, the bd meter is bd's responsibility,
    >>and more

    importantly, that

    >>minimed provides the meter at no additional charge as
    >>part of the

    package. So

    >>much for refunds. And it occured to me that it's still
    >>needed to

    download the

    >>pump to the computer, at least with the data cable
    >>minimed sent me that

    connects

    >>to the test strip port (with great difficulty).
    >
    >Hi Peter,
    >
    >That the meter is BD's responsibility is not something
    >that we can really

    argue.

    >That it came as part of a package, however, is. In fact,
    >I was told quite

    the

    >opposite of what you were told. I was told that the
    >reason for the cost of

    the

    >upgrade was largely because of the purchase that Minimed
    >must make to the

    outside

    >3rd party (BD) for the meter. When I told them I wanted
    >to return the

    entire

    >upgrade, they said NO, you bought it, you own it. This,
    >I thought, was

    reason

    >for me to never consider purchasing product from them
    >again. I took my

    concerns

    >up to a higher level, and was eventually given a refund,
    >and allowed to

    keep the

    >goods.
    >
    >I never got any data cable from Minimed. I use my
    >Glucometer Elite, for

    which

    >I purchased a data cable. If I want to use the bolus
    >wizard, I can simply

    enter

    >the BG value obtained. It saves on the pump battery
    >too. 🙂
    >
    >If you're interested in the email address of the party
    >that ultimately

    helped

    >me out, let me know, and I'll dig it up and post it.
    >
    >Best,
    >
    >Randy
    >


  15. packaging. you can use ultra strips in the ultrasmart meter.

    Marshall wrote:

    All results were pretty much within the same range, though I

    Quoted message said:

    still do not understand why he keeps selling me the Ultra
    Strips, instead of the Ultra Smart Strips.

    Is there a difference?

  16. Bay Area Dave <[email hidden]> wrote in
    :"]news:[email hidden]:

    Quoted message said:

    We will evaluate the other meters on the market, that's
    for sure. I was looking over the Cozmo the other day. My
    doctor has NEVER prescribed it and it looks pretty weird.
    Also, the boluses are given to fast for my taste. What
    else is available in the US that infuses boluses at no
    more than 2 units per minute? Anything??

    I bailed on MiniMed and, in fact, chose the Cozmo. My endo
    currently recommends those two as her preferred choices,
    although she will sign off on any pump you want. I've had it
    for a few months now and am quite happy.

    The extended bolus increments in 15 minute granularity. That
    may go some way to addressing your bolus speed concern. I
    don't know the dose granularity off the top of my head, but
    I'm sure you can get it from Deltec.

    --
    -------
    Charly Coughran [email hidden]

  17. Charly Coughran said:

    Bay Area Dave wrote:

    Quoted message said:
    Quoted message said:

    We will evaluate the other meters on the market, that's
    for sure. I was looking over the Cozmo the other day. My
    doctor has NEVER prescribed it and it looks pretty weird.
    Also, the boluses are given to fast for my taste. What
    else is available in the US that infuses boluses at no
    more than 2 units per minute? Anything??

    Quoted message said:

    I bailed on MiniMed and, in fact, chose the Cozmo. My endo
    currently recommends those two as her preferred choices,
    although she will sign off on any pump you want. I've had
    it for a few months now and am quite happy.

    <snip>

    wow! 🙂

    what will biggs think about this MM revolt?

    snip the above (and i'm guessing that you will), but...

    what made you decide against the Animas pump?

    all ears, bill t1 since '57, ex MiniMed for 10 years

  18. thanks for the info Charly. Yes, 15 minute intervals would
    resolve my concern over fast infusion rates. The MiniMed
    pumps can extend boluses in 30 minute increments. How do you
    like the programming of the Cozmo? Is it straightforward?

    dave

    Charly Coughran said:

    Bay Area Dave <[email hidden]> wrote in
    :"]news:[email hidden]:

    Quoted message said:

    We will evaluate the other meters on the market, that's
    for sure. I was looking over the Cozmo the other day. My
    doctor has NEVER prescribed it and it looks pretty weird.
    Also, the boluses are given to fast for my taste. What
    else is available in the US that infuses boluses at no
    more than 2 units per minute? Anything??

    I bailed on MiniMed and, in fact, chose the Cozmo. My endo
    currently recommends those two as her preferred choices,
    although she will sign off on any pump you want. I've had
    it for a few months now and am quite happy.

    The extended bolus increments in 15 minute granularity.
    That may go some way to addressing your bolus speed
    concern. I don't know the dose granularity off the top of
    my head, but I'm sure you can get it from Deltec.

  19. willbill <[email hidden]> wrote in
    :"]news:[email hidden]:

    Quoted message said:
    Charly Coughran said:

    Bay Area Dave wrote:

    Quoted message said:
    Quoted message said:

    We will evaluate the other meters on the market, that's
    for sure. I was looking over the Cozmo the other day. My
    doctor has NEVER prescribed it and it looks pretty
    weird. Also, the boluses are given to fast for my taste.
    What else is available in the US that infuses boluses at
    no more than 2 units per minute? Anything??

    Quoted message said:

    I bailed on MiniMed and, in fact, chose the Cozmo. My
    endo currently recommends those two as her preferred
    choices, although she will sign off on any pump you want.
    I've had it for a few months now and am quite happy.

    <snip>

    wow! 🙂

    what will biggs think about this MM revolt?

    snip the above (and i'm guessing that you will), but...

    what made you decide against the Animas pump?

    all ears, bill t1 since '57, ex MiniMed for 10 years

    Prejudice mostly. I first encountered Animas at a booth at
    TCOYD when they were brand new. The rep behind the counter
    was blowing smoke like a diesel with no rings and dirty
    injectors. Pure [censored]. Year or so later I listened to a
    woman for an hour after a pumpers club meeting complain
    about how horrible she thought the programming was after
    switching from her MiniMed. I looked at it when I had a
    meeting with my DE to decide and it dropped off the list
    pretty quickly. I forget what specific items in the
    comparison chart turned me off. The DE had one of every
    pump and a nice chart she had made. The decision between
    the MiniMed and the Cozmo was very difficult, but I finally
    decided I'd take the chance rather than deal with the
    Medtronic/MiniMed customer service any more given the
    direction they were going. I do miss the clicking.

    --
    -------
    Charly Coughran [email hidden]

  20. oh, I just remembered why I said the Cozmo looked "weird".
    It's the conical cap where the tubing comes out of the pump.
    It looks overly large in the picture on their website.

    dave

    Charly Coughran said:

    willbill <[email hidden]> wrote in
    :"]news:[email hidden]:

    Quoted message said:
    Charly Coughran said:

    Bay Area Dave wrote:

    Quoted message said:

    >We will evaluate the other meters on the market, that's
    >for sure. I was looking over the Cozmo the other day. My
    >doctor has NEVER prescribed it and it looks pretty
    >weird. Also, the boluses are given to fast for my taste.
    >What else is available in the US that infuses boluses at
    >no more than 2 units per minute? Anything??

    Quoted message said:

    I bailed on MiniMed and, in fact, chose the Cozmo. My
    endo currently recommends those two as her preferred
    choices, although she will sign off on any pump you want.
    I've had it for a few months now and am quite happy.

    <snip>

    wow! 🙂

    what will biggs think about this MM revolt?

    snip the above (and i'm guessing that you will), but...

    what made you decide against the Animas pump?

    all ears, bill t1 since '57, ex MiniMed for 10 years

    Prejudice mostly. I first encountered Animas at a booth at
    TCOYD when they were brand new. The rep behind the counter
    was blowing smoke like a diesel with no rings and dirty
    injectors. Pure [censored]. Year or so later I listened to a
    woman for an hour after a pumpers club meeting complain
    about how horrible she thought the programming was after
    switching from her MiniMed. I looked at it when I had a
    meeting with my DE to decide and it dropped off the list
    pretty quickly. I forget what specific items in the
    comparison chart turned me off. The DE had one of every
    pump and a nice chart she had made. The decision between
    the MiniMed and the Cozmo was very difficult, but I
    finally decided I'd take the chance rather than deal with
    the Medtronic/MiniMed customer service any more given the
    direction they were going. I do miss the clicking.