One Touch Ultra problem?

Started by Kari Karhi · · Last activity · 11 posts · 1,486 views

This thread is locked and is currently read-only.

Thread navigation

Jump through the discussion

Go to the original post, the replies on this page, or the latest preserved contribution.

Thread details

What we know about this thread

Original section
General fitness, health and nutrition
Published
8 May 2004
Last activity
13 June 2004
Original author
Kari Karhi
Posts
11
Discussion status
Public discussion
Total views
1,486
Views / 30 days
0

The navigation and discussion metadata provide context. Posts remain in their original chronological order.

Showing posts 1–11 of 11
Posts remain in their original chronological order.

Text size
  1. Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking a
    reading? Could my tester have a problem? I bought it about
    seven months ago.

  2. I'm also prediabetic, use a one touch ultra and for backup
    the walmart that requires 20 seconds. One touch will vary if
    you use several strips by as much as 5 pts on mine. On
    average they are 8-10 pts higher than Walmart.

    Like everyone, if I don't like the reading I try it again.
    Who said on this board that average reading for over 50
    was 100? What study was that based on? "Kari Karhi"
    <[email hidden]> wrote in message news:eSXmc.11901$V97-
    [email hidden]...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.

  3. Kari Karhi said this...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.


    well within the tolerance of a BG meter.

    In UK units, there is a tolerance of 1 unit, which is 18 of
    yours. It's a little bit of plastic you shove in your
    pocket, not a portable lab after all!

    If in doubt, ring Lifescan and get some test fluid (free
    of charge)

    R
    --
    Now with 64 bit stupidity

    www.flyingrat.net

  4. Flying Rat said:

    Kari Karhi said this...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.

    well within the tolerance of a BG meter.

    In UK units, there is a tolerance of 1 unit, which is 18
    of yours. It's a little bit of plastic you shove in your
    pocket, not a portable lab after all!

    If in doubt, ring Lifescan and get some test fluid (free
    of charge)

    R


    Hmm, that not my experience. Overall, my readings vary by
    not more than 5 US units. If I repeat readings, I have only
    a +-1 unit variation. This ignores those days when I have a
    late-night sweet snack, with full knowledge of what it means
    the next morning. Looking back on six months of data, I can
    clearly see a trend of readings starting at 110 average to
    the present average of 105, gotten by losing 30 lb. I figure
    another 20 lb more should take me to around a 100 average,
    which I can live with.

  5. beachnut said:

    I'm also prediabetic, use a one touch ultra and for backup
    the walmart that requires 20 seconds. One touch will vary
    if you use several strips by as much as 5 pts on mine. On
    average they are 8-10 pts higher than Walmart.

    Like everyone, if I don't like the reading I try it
    again. Who said on this board that average reading for
    over 50 was 100? What study was that based on? "Kari
    Karhi" <[email hidden]> wrote in message news:eSXmc.-
    [email hidden]...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.


    So if you are right, since I am in mid-fifties, my One Touch
    readings should be in the 110 range and I am not pre-
    diabetic after all! Please don't tell me that, I have lost
    30 lb so far with another 20 lb to go. I would never have
    gone on a diet if it wasn't for this scare. Best thing that
    could have happened to me.

  6. Kari Karhi said this...

    Quoted message said:

    Hmm, that not my experience.


    so your 'experience' is more valid than Lifescan's own
    product data?

    R
    --
    Now with 64 bit stupidity

    www.flyingrat.net

  7. Kari Karhi said this...

    Quoted message said:

    Hmm, that not my experience.


    My reply was taken directly from Lifescan published data on
    One Touch Ultra meters.

    R
    --
    Now with 64 bit stupidity

    www.flyingrat.net

  8. I also use the same meter. I use to worry about the readings
    bouncing around like the ones you are getting. The tolerance
    for the ultra measurements are plus or minus 20% compared to
    a clinal lab measurement. The ultra is susposed to give you
    that reading and that is within federal guidelines. I always
    test before a ceratain meal and never test after a meal til
    2 hrs after a meal. If you test before it can throw off your
    measurements. Accu-scan tolerance is not much better (+/- 15
    to 20 %). Try this. Take your measurements over a certain
    length of time and plot a graph of the data. Your BGL's are
    everywhere on the graph (hopefully most are within the
    target range (mine is 90-140 overall). Take the average for
    each days BGL and then plot it. Hopefully you should be near
    your target range. For the month of may, the average daily
    glucose levels were within the 90 - 140 target range. I am
    type 2. My glucose readings are an average of 3.6 times a
    day. Look at the container of strips. The control range is
    92-124 mg/dL. Do a control test and if it is within the
    control range, you are okay. remember, you will get
    different readings with every strip. Say, you had enough
    blood on your finger and you had two meters with strips
    inserted ready to take a reading and you dipped both strips
    in the test blood at the same time, you will get 2 different
    readings. The strips and meter work on a electric current
    when trying to determine your BGL. It aint going to happen
    trying to get the same BGL out of the 2 strips using the
    same test conditions. Plus, everything I read on the ultra
    says it is very accurate (remember +/- 20%) and comes highly
    recommended.

    "Kari Karhi" <[email hidden]> wrote in message news:eSXm-
    [email hidden]...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.

  9. I read the docs and as you said, there is the 20% uncertainty from
    the lab reading. I think it probably is smaller but the company
    is leaving some wiggle room there. However, I am mostly interested
    in the consistency of the readings, which is a somewhat different
    matter. My readings are pretty consistent, only varying by about
    two or three units from day to day. If I believed the 20%, my
    reading could be 80 one day and 120 the next. I do not see that
    at all. I think that the FBG readings should be consistent, if
    you really fast at least 8 hours and have not had too much carbs
    the previous day. I only take one reading a day, so I can not
    average them. I'm going to stop retesting and just live with the
    anomalous high readings. Unfortunately, I never get anomalous
    low readings :-(

    Charles Eaves said:

    I also use the same meter. I use to worry about the
    readings bouncing around like the ones you are getting.
    The tolerance for the ultra measurements are plus or minus
    20% compared to a clinal lab measurement. The ultra is
    susposed to give you that reading and that is within
    federal guidelines. I always test before a ceratain meal
    and never test after a meal til 2 hrs after a meal. If you
    test before it can throw off your measurements. Accu-scan
    tolerance is not much better (+/- 15 to 20 %). Try this.
    Take your measurements over a certain length of time and
    plot a graph of the data. Your BGL's are everywhere on the
    graph (hopefully most are within the target range (mine is
    90-140 overall). Take the average for each days BGL and
    then plot it. Hopefully you should be near your target
    range. For the month of may, the average daily glucose
    levels were within the 90 - 140 target range. I am type 2.
    My glucose readings are an average of 3.6 times a day.
    Look at the container of strips. The control range is 92-
    124 mg/dL. Do a control test and if it is within the
    control range, you are okay. remember, you will get
    different readings with every strip. Say, you had enough
    blood on your finger and you had two meters with strips
    inserted ready to take a reading and you dipped both
    strips in the test blood at the same time, you will get 2
    different readings. The strips and meter work on a
    electric current when trying to determine your BGL. It
    aint going to happen trying to get the same BGL out of the
    2 strips using the same test conditions. Plus, everything
    I read on the ultra says it is very accurate (remember +/-
    20%) and comes highly recommended.

    "Kari Karhi" <[email hidden]> wrote in message news:eS-
    [email hidden]...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using the
    One Touch Ultra tester. I've noticed that over all my
    readings are pretty consistent, but there are exceptions.
    Occasionally, maybe once or twice a month, my reading is
    much higher. Today, for example, my reading was 109, when
    it usually is under 105. I took another strip and used the
    same puncture to get another reading of 99. Then, to make
    sure, I had another puncture and that reading was 101,
    which I recorded. I thought at one time that maybe I just
    let some air into the strip, but I've been pretty careful
    recently and today I was positive there was no air in the
    sample. Can the samples really vary that much and should I
    really be taking several measuments every day and average
    them? Or am I doing something wrong? Do people wipe the
    first drop away and squeeze some more blood before taking
    a reading? Could my tester have a problem? I bought it
    about seven months ago.


  10. Try two readings a day for a while. Maybe a week. Download
    the lifescan 2.1 program. It is free. Enter your data. Look
    at the reports. Look at the summary and standard report. All
    reports in the program are great. If your levels are like
    mine, most readings are inside the target rangebut there are
    readings outside the target level . See standard report.

    Now look at the summary report. especially the graph
    (average glucose vs meal type).

    "Kari Karhi" <[email hidden]> wrote in message news:8tSw-
    [email hidden]...

    Quoted message said:

    I read the docs and as you said, there is the 20%
    uncertainty from the lab reading. I think it probably is
    smaller but the company is leaving some wiggle room there.
    However, I am mostly interested in the consistency of the
    readings, which is a somewhat different matter. My
    readings are pretty consistent, only varying by about two
    or three units from day to day. If I believed the 20%, my
    reading could be 80 one day and 120 the next. I do not see
    that at all. I think that the FBG readings should be
    consistent, if you really fast at least 8 hours and have
    not had too much carbs the previous day. I only take one
    reading a day, so I can not average them. I'm going to
    stop retesting and just live with the anomalous high
    readings. Unfortunately, I never get anomalous low
    readings :-(

    Charles Eaves said:

    I also use the same meter. I use to worry about the
    readings bouncing


    around

    Quoted message said:
    Quoted message said:

    like the ones you are getting. The tolerance for the
    ultra measurements


    are

    Quoted message said:
    Quoted message said:

    plus or minus 20% compared to a clinal lab measurement.
    The ultra is susposed to give you that reading and that
    is within federal


    guidelines. I

    Quoted message said:
    Quoted message said:

    always test before a ceratain meal and never test after
    a meal til 2 hrs after a meal. If you test before it can
    throw off your measurements. Accu-scan tolerance is not
    much better (+/- 15 to 20 %). Try this. Take your
    measurements over a certain length of time and plot


    a

    Quoted message said:
    Quoted message said:

    graph of the data. Your BGL's are everywhere on the
    graph (hopefully


    most

    Quoted message said:
    Quoted message said:

    are within the target range (mine is 90-140 overall).
    Take the average for each days BGL and then plot it.
    Hopefully you


    should be

    Quoted message said:
    Quoted message said:

    near your target range. For the month of may, the
    average daily glucose levels were within the 90 - 140
    target range. I am type 2. My glucose readings are an
    average of 3.6 times a day. Look at the container of
    strips. The control range is 92-124 mg/dL. Do a control
    test and if it is within the control range, you are
    okay. remember, you will get different readings with
    every strip. Say, you had enough blood on your finger
    and you had two meters with strips inserted ready to
    take a reading and you dipped both strips in the test
    blood at


    the

    Quoted message said:
    Quoted message said:

    same time, you will get 2 different readings. The strips
    and meter work


    on a

    Quoted message said:
    Quoted message said:

    electric current when trying to determine your BGL. It
    aint going to


    happen

    Quoted message said:
    Quoted message said:

    trying to get the same BGL out of the 2 strips using the
    same test conditions. Plus, everything I read on the
    ultra says it is very accurate (remember


    +/-

    Quoted message said:
    Quoted message said:

    20%) and comes highly recommended.

    "Kari Karhi" <[email hidden]> wrote in message news:-
    [email hidden]...

    Quoted message said:

    Hi, I'm pre-diabetic and take my FBG every day, using
    the One Touch Ultra tester. I've noticed that over all
    my readings are pretty consistent, but there are
    exceptions. Occasionally, maybe once or twice a month,
    my reading is much higher. Today, for example, my
    reading was 109, when it usually is under 105. I took
    another strip and used the same puncture to get another
    reading of 99. Then, to make sure, I had another
    puncture and that reading was 101, which I recorded. I
    thought at one time that maybe I just let some air into
    the strip, but I've been pretty careful recently and
    today I was positive there was no air in the sample. Can
    the samples really vary that much and should I really be
    taking several measuments every day and average them? Or
    am I doing something wrong? Do people wipe the first
    drop away and squeeze some more blood before taking a
    reading? Could my tester have a problem? I bought it
    about seven months ago.

  11. I might give two readings a day a try, good suggestion. I
    have the program, right from the beginning so I have about
    seven months of data. I have been above target six times,
    most at the beginning. I enter the data manually as I did
    not see the need to buy the data cable. The reports are ok,
    but I wish one could change the scale and it could have some
    horizontal stripes so it would be easier to read. Are people
    out there using other programs with the One Touch meter?

    Charles Eaves said:

    Try two readings a day for a while. Maybe a week. Download
    the lifescan 2.1 program. It is free. Enter your data.
    Look at the reports. Look at the summary and standard
    report. All reports in the program are great. If your
    levels are like mine, most readings are inside the target
    rangebut there are readings outside the target level . See
    standard report.

    Now look at the summary report. especially the graph
    (average glucose vs meal type).

    "Kari Karhi" <[email hidden]> wrote in message news:8t-
    [email hidden]...

    Quoted message said:

    I read the docs and as you said, there is the 20%
    uncertainty from the lab reading. I think it probably is
    smaller but the company is leaving some wiggle room there.
    However, I am mostly interested in the consistency of the
    readings, which is a somewhat different matter. My
    readings are pretty consistent, only varying by about two
    or three units from day to day. If I believed the 20%, my
    reading could be 80 one day and 120 the next. I do not see
    that at all. I think that the FBG readings should be
    consistent, if you really fast at least 8 hours and have
    not had too much carbs the previous day. I only take one
    reading a day, so I can not average them. I'm going to
    stop retesting and just live with the anomalous high
    readings. Unfortunately, I never get anomalous low
    readings :-(

    Charles Eaves said:

    I also use the same meter. I use to worry about the
    readings bouncing

    around

    Quoted message said:
    Quoted message said:

    like the ones you are getting. The tolerance for the
    ultra measurements

    are

    Quoted message said:
    Quoted message said:

    plus or minus 20% compared to a clinal lab measurement.
    The ultra is susposed to give you that reading and that
    is within federal

    guidelines. I

    Quoted message said:
    Quoted message said:

    always test before a ceratain meal and never test after a
    meal til 2 hrs after a meal. If you test before it can
    throw off your measurements. Accu-scan tolerance is not
    much better (+/- 15 to 20 %). Try this. Take your
    measurements over a certain length of time and plot

    a

    Quoted message said:
    Quoted message said:

    graph of the data. Your BGL's are everywhere on the graph
    (hopefully

    most

    Quoted message said:
    Quoted message said:

    are within the target range (mine is 90-140 overall).
    Take the average for each days BGL and then plot it.
    Hopefully you

    should be

    Quoted message said:
    Quoted message said:

    near your target range. For the month of may, the average
    daily glucose levels were within the 90 - 140 target
    range. I am type 2. My glucose readings are an average of
    3.6 times a day. Look at the container of strips. The
    control range is 92-124 mg/dL. Do a control test and if
    it is within the control range, you are okay. remember,
    you will get different readings with every strip. Say,
    you had enough blood on your finger and you had two
    meters with strips inserted ready to take a reading and
    you dipped both strips in the test blood at

    the

    Quoted message said:
    Quoted message said:

    same time, you will get 2 different readings. The strips
    and meter work

    on a

    Quoted message said:
    Quoted message said:

    electric current when trying to determine your BGL. It
    aint going to

    happen

    Quoted message said:
    Quoted message said:

    trying to get the same BGL out of the 2 strips using the
    same test conditions. Plus, everything I read on the
    ultra says it is very accurate (remember

    +/-

    Quoted message said:
    Quoted message said:

    20%) and comes highly recommended.

    "Kari Karhi" <[email hidden]> wrote in message news:e-
    [email hidden]...

    >Hi, I'm pre-diabetic and take my FBG every day, using
    >the One Touch Ultra tester. I've noticed that over all
    >my readings are pretty consistent, but there are
    >exceptions. Occasionally, maybe once or twice a month,
    >my reading is much higher. Today, for example, my
    >reading was 109, when it usually is under 105. I took
    >another strip and used the same puncture to get another
    >reading of 99. Then, to make sure, I had another
    >puncture and that reading was 101, which I recorded. I
    >thought at one time that maybe I just let some air into
    >the strip, but I've been pretty careful recently and
    >today I was positive there was no air in the sample. Can
    >the samples really vary that much and should I really be
    >taking several measuments every day and average them? Or
    >am I doing something wrong? Do people wipe the first
    >drop away and squeeze some more blood before taking a
    >reading? Could my tester have a problem? I bought it
    >about seven months ago.